| Recorded input | Required observation | Baseline observation | Corrected observation |
|---|---|---|---|
Missing input (T07): None
|
state: unavailable; text: Unavailable; unit: none |
state: value; text: 0.0; unit: u
|
state: unavailable; text: Unavailable; unit: none |
Valid zero (T01): exact integer 0
|
state: value; text: 0.0; unit: u
|
state: value; text: 0.0; unit: u
|
state: value; text: 0.0; unit: u
|
This is observed function output, not a device screenshot. The arbitrary unit u and input ticks represent no clinical measurement or threshold. Inspect the baseline and corrected output.
01 / Establish the basis
What IEC 62304 covers, and who decides what applies
IEC 62304 provides a software lifecycle framework for medical-device software, including software embedded in a device and software that is itself a device. Its published scope excludes validation and final release of the medical device itself, including a software-only device. That device-level exclusion does not remove software release activities or make validation unnecessary. Keep those responsibilities identified in the governing plans. IEC
Start the change review by separating five sources of authority:
| Basis | What to record | What it establishes |
|---|---|---|
| Publisher edition | IEC 62304:2006+AMD1:2015 CSV, edition 1.1, published 26 June 2015 | The consolidated first edition and Amendment 1. Here CSV means consolidated version. |
| National adoption | The exact adopted designation and amendment | A national publication identity, which can have a different year. |
| Regulator recognition or guidance | Jurisdiction, record, edition, extent and applicable guidance | The particular regulator's treatment of a source. |
| Project adoption | Approved plan, revision, applicable sources, rationale and decision owner | The actual basis governing this project's work. |
| Engineering advice | A method or template chosen to help the review | A useful working approach whose suitability still needs assessment. |
For example, FDA recognition record 13-79 lists IEC edition 1.1, 2015-06, and the identical adoption ANSI/AAMI IEC 62304:2006/A1:2016, with complete-standard recognition. The national adoption's 2016 label is not a new IEC edition. Recognition does not approve a device or tool, and recognition of later revisions is not automatic.
The IEC record's 2028 stability date concerns standards maintenance; it is neither expiry nor a project transition deadline. The IEC stability explanation and April 2026 TC 62 strategic plan have different roles. The latter places a second edition in future work with a 2028 planning target. It does not establish a published second edition or promise a publication date.
Keep risk, documentation and validation decisions distinct
The reviewer needs the project's software safety classification, rationale and responsible authority. FDA's Basic and Enhanced Documentation Levels describe recommended submission documentation. They are not IEC software safety classes or device regulatory classes. FDA software guidance, sections IV–V
ISO 14971:2019, third edition, confirmed in 2025, addresses medical-device risk management. Its current status does not silently update a normative reference in another adopted standard. Historical IEC/TR 80002-1:2009 explains software risk management using ISO 14971:2007 and IEC 62304:2006; it adds no requirements to those editions and is not an inspection or certification basis.
For US work, FDA states that QMSR became effective on 2 February 2026. Older guidance's QSR citations are not a complete current legal checklist. Ask the identified quality/regulatory authority to resolve the actual basis before making a conformity or release claim. The worksheet below organizes one review; it does not determine these decisions.
02 / Start with the record
Free, complete change-impact and evidence worksheet
Copy the whole record below. Reference existing controlled plans instead of duplicating them. Repeat the contract, impact, campaign and anomaly rows as needed. Use explicit values such as unknown, not performed or awaiting named authority, with an owner and resolution trigger. An empty cell must not look like a completed review.
This is an original engineering worksheet. It is useful without Proof and is not a clause checklist or a prescribed document count.
CHANGE-IMPACT AND EVIDENCE RECORD
1. CHANGE AND AUTHORITY
Record ID / revision / preparer / status:
Product or software item / intended use / review boundary:
Change reason / requested decision / decision criteria:
Previous identity / candidate identity / source and build configuration:
Adopted lifecycle and change plans, revisions and approval references:
Standards, editions and amendments; governing or informative role:
National adoption / market recognition, edition and extent:
Software safety class / rationale reference / deciding authority:
FDA documentation level, if relevant / separate rationale and authority:
Requirement owner / risk owner / quality reviewer / release authority:
Required role separation / actual assignment / unresolved authority:
2. BEHAVIORAL CONTRACT (repeat per changed or affected requirement)
Requirement ID / revision / source / authority:
Previous behavior / intended behavior / intentional difference:
Input types and domain / units / conversion / precision / rounding:
Valid zero / missing / invalid / stale-data behavior:
Observable outputs, state, labels, side effects and error reasons:
Expected observations / independent oracle and its justification:
Unresolved behavior / decision owner / resolution trigger:
3. RISK-RELATED AND INTERFACE IMPACT (repeat per relationship)
Risk or control record / revision / associated requirement:
Relationship meaning / affected failure behavior / open assumptions:
Affected design, code, producer, consumer, interface and configuration:
Dependency role / version / configuration / known problems / maintenance:
Applicable OTS or SOUP assessment reference and responsible authority:
Evidence to renew / evidence proposed for reuse / reason for each:
Excluded work / accountable plan or owner / required completion point:
4. EVIDENCE CAMPAIGNS (repeat per method and claim)
Campaign ID / requirement revision / method / executor and reviewer:
Inputs and preconditions / expected observation / oracle reference:
Counting unit and scope / overlaps / excluded states:
Candidate, source, input, runner and procedure identities:
Target, environment, tools, versions and configuration:
Execution or review time / command or procedure / raw result location:
Observed result / failed IDs / incomplete or blocked work:
Evidence status / currency or reuse rationale / approving authority:
5. ANOMALIES AND BOUNDED DECISION
Anomaly ID / affected basis / expected versus observed / discovery:
Cause assessment / correction / retained failure / rerun evidence:
Disposition / rationale / owner / review record / revisit trigger:
Unperformed work and blockers / owners / completion conditions:
Conclusion supported by current evidence / explicit limits:
Quality review: reviewer / record / decision / conditions:
Software release decision and configuration reference:
Separate device validation and final-release decision reference:
Reassessment triggers / next responsible person and action:
Keep three questions visible as you fill it: Was the work performed? Does its basis still match? Who accepted its use for this decision? A result can be executed yet stale, or current yet awaiting review. An approved exception needs its own rationale and scope; it does not change a failed observation into a passing one.
03 / Follow an actual failure
A display conversion is corrected and reexecuted
This original nonclinical Python fixture makes the display-conversion example inspectable. Its fictional owner states the contract in requirements.json. No clinical thresholds, medical performance or real device are modeled.
The contract comes before the implementation
Available inputs are exact integers from 0 through 40 inclusive. Four ticks equal one arbitrary u. Display one decimal place; at an exact nonnegative halfway value, round upward to the next tenth. Thus one tick is 0.25 u and displays 0.3.
None means unavailable: return state=unavailable, text Unavailable and no unit. Integer zero remains a valid value: state=value, text 0.0, unit u.
All negative integers and integers above 40 are invalid with reason RANGE. Apart from None, booleans, strings, floats and all other inputs whose type is not exactly int are invalid with reason TYPE. Invalid output has state=invalid, text Invalid input and no unit. Negative inputs never reach rounding.
The baseline substitutes zero for None and uses ties-to-even formatting. Both choices contradict this contract. They are not inherently wrong policies for every application; the defect is disagreement with the stated requirement.
Twelve selected observations
In the table, numeric text means state=value with unit u; unavailable and invalid outputs have no unit. The raw files preserve every field.
| Case | Input | Expected | Baseline | Corrected |
|---|---|---|---|---|
| T01 |
0
|
0.0
|
Matches | Matches |
| T02 |
4
|
1.0
|
Matches | Matches |
| T03 |
40
|
10.0
|
Matches | Matches |
| T04 |
1
|
0.3
|
0.2, fails
|
Matches |
| T05 |
3
|
0.8
|
Matches | Matches |
| T06 |
5
|
1.3
|
1.2, fails
|
Matches |
| T07 |
None
|
Unavailable state and text |
Value 0.0 u, fails
|
Matches |
| T08 |
-1
|
Invalid, RANGE
|
Matches | Matches |
| T09 |
41
|
Invalid, RANGE
|
Matches | Matches |
| T10 | string "4"
|
Invalid, TYPE
|
Matches | Matches |
| T11 | Boolean True
|
Invalid, TYPE
|
Matches | Matches |
| T12 | float 4.0
|
Invalid, TYPE
|
Matches | Matches |
The rounding failure is selective. T05 already passes because 0.75 becomes 0.8 under both policies. A single convenient halfway example would miss the disagreement seen at 0.25 and 1.25.
The implemented correction and independent expectation
The corrected implementation handles missing input separately, validates exact type and range, then computes tenths with integer arithmetic:
tenths = (raw * 10 + 2) // 4
text = f"{tenths // 10}.{tenths % 10}"
This excerpt is the conversion after validation, not a standalone input handler. Inspect the complete display.py. All four implementation modes live in this one frozen source file. The recorded mode and shared source hash identify each variant.
Expected results are literal data in cases.json and the 41-row integer oracle. The runner independently cross-checks every oracle row using Decimal(raw) / Decimal(4) and ROUND_HALF_UP. Python's Decimal documentation defines the arithmetic and rounding operations.
That second calculation challenges transcription mistakes without reusing the correction's integer formula. More fundamentally, the expected behavior was stated independently of the implementation's output. Code and tests can agree on the same wrong rounding policy if both derive from the old function. Computational diversity helps check the chosen contract; it does not approve the contract or establish independent personnel review.
What the final campaign actually recorded
| Implementation | Selected checks passed | Integer-domain checks passed | Selected failures | Child exit |
|---|---|---|---|---|
baseline
|
9/12 | 31/41 | T04, T06, T07 | 1 |
corrected
|
12/12 | 41/41 | None | 0 |
mutant_ties_even
|
10/12 | 31/41 | T04, T06 | 1 |
mutant_missing_as_zero
|
11/12 | 41/41 | T07 | 1 |
These are separate, overlapping campaigns. The integer campaign enumerates the entire accepted integer domain, while the selected campaign also examines missing and particular invalid inputs. Do not add their denominators or turn them into a conformity percentage.
The baseline's integer failures occur at 1, 5, 9, 13, 17, 21, 25, 29, 33, 37: each expected text ending .3 instead ends .2. The corrected implementation passes all 41 rows.
The two hand-applied mutants isolate the defects. The missing-as-zero mutant passes 41/41 integers while failing T07. Exhaustive evidence for the conversion domain leaves the unavailable-data promise unanswered. No external mutation engine ran, and these two fault checks do not establish a general test-adequacy score.
The final public capture started 27 October 2026 at 22:04:12.790055 UTC and completed at 22:04:12.957248 UTC, using CPython 3.14.7 on macOS-26.5-arm64-arm-64bit-Mach-O. The outer runner returned 0 because the expected baseline/mutant failures and corrected passes reproduced. It did not accept a device or release.
Inspect baseline output · Inspect corrected output · Download and reproduce the evidence kit · Commands and hashes
Filled change-impact and evidence record
The following is a fictional change-review record filled from those actual observations. Unassigned device responsibilities remain explicit. Relative file names resolve inside the linked evidence kit.
CHANGE-IMPACT AND EVIDENCE RECORD
1. CHANGE AND AUTHORITY
Record: DISPLAY-CHANGE-01 / revision 1 / teaching record / observed correction.
Preparer: this guide's teaching-record author; no project approval asserted.
Evidence root: https://reqproof.com/assets/iec-62304/research-rerun/
Item and intended use: fictional text/state converter; nonclinical education.
Boundary: function output, accepted integer domain and selected other inputs.
Reason: baseline violates declared rounding and unavailable-state behavior.
Requested decision: does corrected mode meet this campaign's contract?
Criteria: match all selected records and all 41 integer oracle rows.
Previous/candidate: baseline/corrected modes in the same fixture/display.py.
Configuration: mode selection; frozen source and inputs in captured/provenance.json.
Adopted lifecycle/change plan: no device-project plan or approval exists here.
Authored teaching contract: fixture/requirements.json.
Standards: IEC/ISO/FDA references are informative subject context here.
National adoption/recognition: no device-project adoption or submission claimed.
IEC software safety class/rationale/authority: not assigned; no assessment.
FDA documentation level/rationale/authority: not assigned.
Requirement owner: fictional fixture owner identified by the contract.
Risk owner, quality reviewer, release authority: no device roles assigned.
Role separation: no independent personnel review or project authorization claimed.
2. BEHAVIORAL CONTRACT
Requirement identities: DISP-01 conversion/rounding; DISP-02 unavailable;
DISP-03/DISP-04 input rejection and output representation.
Source: fixture/requirements.json; frozen SHA-256:
1dbf9c38050fcae55ac1ba613fe76071aacd0fd8d72c4d95e6894d4182140a1e
Previous: ties-even text; missing input substituted with zero.
Intended difference: nonnegative halfway values round up; missing stays unavailable.
Domain: exact integers 0..40; four ticks per u; one decimal place.
Zero: value / 0.0 / u. Missing None: unavailable / Unavailable / no unit.
Invalid integer range: invalid / Invalid input / no unit / RANGE.
Invalid type, including bool and float: same invalid representation / TYPE.
Stale data: no contract or execution in this fixture.
Outputs: state, text, unit and invalid reason; no device-side effects assessed.
Oracle: literal fixture/cases.json and fixture/integer-domain-oracle.json expectations;
41 integer rows cross-checked using Decimal and ROUND_HALF_UP.
Unresolved behavior: real use policy and stale-state behavior are outside scope;
an actual requirement authority must resolve them before any product claim.
3. RISK-RELATED AND INTERFACE IMPACT
Risk/control records: no actual medical-device risk records exist in this fixture.
Review concern: a plausible numeric output can hide a rounding or missing-state error.
Affected path: raw input -> selected converter -> state/text/unit output.
Affected implementation: missing handling and numerical representation.
Dependencies: CPython 3.14.7 standard library; Decimal used for oracle checking.
Known-problem/maintenance suitability and OTS/SOUP assessment: not performed.
Renewed evidence: selected contract checks and full accepted integer campaign.
Reuse: baseline retained as failure evidence; it is not candidate passing evidence.
Excluded: real producers/consumers, UI, hardware and clinical use.
Owners/completion point: actual project owners must be assigned before extending scope.
4. EVIDENCE CAMPAIGNS
Methods: exact output comparisons; isolated hand-applied fault contrasts.
Executor: captured reproduce.py and child runner; no named quality reviewer.
Preconditions: extracted kit root, frozen files, implementation selected explicitly.
Selected scope: T01..T12; includes valid, missing and selected invalid inputs.
Integer scope: I00..I40; exact integers 0..40; overlaps selected valid cases.
Claim links: DISP-01 uses integer and selected conversion cases;
DISP-02 uses T07; DISP-03/DISP-04 use the selected full-output comparisons.
Expected: literal full output objects; integer oracle cross-check passes 41 rows.
Identity authority: captured/provenance.json source_sha256 and run command arrays.
Runtime: CPython 3.14.7; macOS-26.5-arm64-arm-64bit-Mach-O.
Start UTC: 2026-10-27T22:04:12.790055+00:00.
End UTC: 2026-10-27T22:04:12.957248+00:00.
Portable reproduction: python3 -B reproduce.py --output rerun-results
Raw observations: captured/<implementation>-stdout.json; stderr files also retained.
baseline: 9/12 selected, 31/41 integers; T04/T06/T07 fail; child exit 1.
corrected: 12/12 selected, 41/41 integers; no failed cases; child exit 0.
mutant_ties_even: 10/12 selected, 31/41 integers; T04/T06 fail; exit 1.
mutant_missing_as_zero: 11/12 selected, 41/41 integers; T07 fails; exit 1.
Outer exit: 0, expected contrast reproduced.
Currency: retained results support this frozen basis; no other target reuse approved.
5. ANOMALIES AND BOUNDED DECISION
Teaching anomaly A1: T04/T06 and ten integer rows disagree with rounding contract.
Cause/correction: ties-even replaced by integer half-up computation after validation.
Teaching anomaly A2: T07 substitutes numeric zero for unavailable input.
Cause/correction: None handled separately before conversion.
Failure/rerun: baseline and corrected raw outputs retained; isolated mutants detected.
Disposition: both defects corrected within the recorded campaign;
no medical-device anomaly closure or quality acceptance has been recorded.
Unperformed: device/UI integration, clinical evaluation, risk/class assignment,
timing, stale-data and concurrency work, Proof audit/coverage measurement,
complete lifecycle review, device validation and release review.
Conclusion: corrected mode matches the fictional contract's recorded checks.
Quality review/reviewer/conditions: no device-project decision or reviewer assigned.
Software release configuration/decision: none.
Device validation/final-release decision: none.
Reassess if contract, source, inputs, runner, runtime or intended context changes.
Next action: a real project assigns authorities and plans evidence for its actual scope.
The correction succeeds within the recorded fictional contract and campaign. Real integration, risk, validation and release activities remain unperformed. A project applying this approach needs its own approved behavior, assessment and evidence.
Follow the change through six review steps
Download the complete video transcript.
Read the complete video transcript
This recorded example is a fictional text converter, with no clinical thresholds or real device. Only exact integers from zero through forty are available values. Four ticks equal one arbitrary u. The contract requires one decimal place and upward rounding at nonnegative halfway values. Missing input, represented by None, must return state unavailable, text Unavailable and no unit; valid zero remains zero point zero with unit u.
The baseline passes nine of twelve selected checks and thirty-one of forty-one integer checks. One tick displays zero point two instead of zero point three. Five ticks display one point two instead of one point three. Missing input becomes a numeric zero. These are retained function observations, not a device display.
The implemented correction handles None separately, rejects invalid types and ranges, then calculates display tenths using integer arithmetic. Negative inputs never reach rounding. Valid zero remains zero point zero with unit u. The reexecuted correction passes twelve of twelve selected checks and all forty-one accepted integers on the captured basis.
Expected outputs are literal records, stated independently of the converter. The runner cross-checks all forty-one integer expectations with Decimal and an explicit halfway rounding mode. This challenges the answer table without copying the correction's formula. It does not establish that the fictional policy suits a medical use, or that an independent quality review occurred.
Each fault was deliberately restored on its own. Ties-even rounding is detected: ten of twelve selected checks and thirty-one of forty-one integers pass. Missing-as-zero passes eleven of twelve selected checks and all forty-one integers, yet violates the separate missing-input promise. The campaigns overlap. Integer checks alone miss that fault; reproducing the expected contrast does not grant release acceptance.
For your own change, identify the adopted plan, requirement authority, risk-related links, affected configurations and quality reviewer. Decide which evidence needs renewal and preserve unresolved anomalies. That is proposed work. This fixture performed no device integration, clinical evaluation, classification, timing, stale-data or concurrency assessment, Proof measurement, complete lifecycle review or release decision.
Read the text walkthrough
1. Establish the contract
This recorded example is a fictional text converter, with no clinical thresholds or real device. Only exact integers from zero through forty are available values. Four ticks equal one arbitrary u. The contract requires one decimal place and upward rounding at nonnegative halfway values. Missing input, represented by None, must return state unavailable, text Unavailable and no unit; valid zero remains zero point zero with unit u.
2. Inspect the baseline
The baseline passes nine of twelve selected checks and thirty-one of forty-one integer checks. One tick displays zero point two instead of zero point three. Five ticks display one point two instead of one point three. Missing input becomes a numeric zero. These are retained function observations, not a device display.
3. Inspect the correction
The implemented correction handles None separately, rejects invalid types and ranges, then calculates display tenths using integer arithmetic. Negative inputs never reach rounding. Valid zero remains zero point zero with unit u. The reexecuted correction passes twelve of twelve selected checks and all forty-one accepted integers on the captured basis.
4. Challenge the expectation
Expected outputs are literal records, stated independently of the converter. The runner cross-checks all forty-one integer expectations with Decimal and an explicit halfway rounding mode. This challenges the answer table without copying the correction's formula. It does not establish that the fictional policy suits a medical use, or that an independent quality review occurred.
5. Keep the campaigns separate
Each fault was deliberately restored on its own. Ties-even rounding is detected: ten of twelve selected checks and thirty-one of forty-one integers pass. Missing-as-zero passes eleven of twelve selected checks and all forty-one integers, yet violates the separate missing-input promise. The campaigns overlap. Integer checks alone miss that fault; reproducing the expected contrast does not grant release acceptance.
6. Apply the record to a real change
For your own change, identify the adopted plan, requirement authority, risk-related links, affected configurations and quality reviewer. Decide which evidence needs renewal and preserve unresolved anomalies. That is proposed work. This fixture performed no device integration, clinical evaluation, classification, timing, stale-data or concurrency assessment, Proof measurement, complete lifecycle review or release decision.
04 / Work in decision order
How to review one software change
Use this procedure to populate the worksheet. It is an engineering workflow for a bounded review, not a reproduction of licensed IEC clauses.
-
Fix the decision and authority. Name the software item, previous and candidate identities, adopted plan and requested decision. Identify who owns requirements, risk interpretation, quality review and release. If the basis is unresolved, assign that decision; do not guess a class from code size or a passing suite.
-
State what changes and what must remain true. Preserve the source and authority behind each affected behavior. Specify types, units, domain, rounding, missing and invalid states, interfaces and side effects. Distinguish an intentional change from a regression. In the fixture, replacing ties-even is intentional because the stated contract already requires halfway-up.
-
Follow risk-related and interface relationships. Start from the changed behavior, then examine relevant risk controls, requirements, design elements, producers and consumers. Explain each relationship: which output could become wrong, who consumes it, and what assumption might fail? A linked hazard name alone does not answer those questions.
-
Choose new evidence and justify reuse. For each affected claim, specify the observation, method, environment and owner. Separate conversion checks from acquisition, stale-state, UI and intended-use questions. Reuse a prior result only with an explicit assessment of the requirement, implementation, configuration, dependency and environment differences.
-
Execute against identified inputs and retain failures. Record raw results, failed IDs, runtime and source/test identities. Investigate failures before changing the expectation. Preserve why a correction addresses the observed mechanism, then rerun affected checks. The original failure remains useful after repair.
-
Review regression and unresolved anomalies. Examine related paths beyond the reproducer, including unchanged consumers and configurations. Record each remaining anomaly's impact assessment, owner and authorized disposition. Missing evidence is an open task; a known defect is an observed problem. Keep their states distinguishable.
-
Record the bounded decision and handoff. State exactly what the current evidence supports, which revisions it covers, and what remains required by other plans. The designated reviewer records the quality decision. Preserve separate software release controls and the device-level validation/final-release decision where applicable.
FDA's software guidance supports retaining tested versions, intentional corrections, regression analysis, results and unresolved anomalies. These are documentation recommendations in sections VI.H–J, rather than a claim that this seven-step procedure is the standard's prescribed process.
05 / Make gaps visible
Change-evidence review checklist
Use this as a conversation with the responsible reviewer. Mark a gap with its owner and next action instead of calculating an aggregate readiness percentage.
| Review question | Evidence to inspect | If absent |
|---|---|---|
| Is the governing basis identifiable? | Plan revision, adopted editions, applicability and authority | Resolve the basis before claiming conformity. |
| Is the intended behavior precise? | Requirement, states, units, domain and oracle | Return the ambiguity to its owner. |
| Are affected relationships explained? | Risk/control links, interfaces, consumers and assumptions | Review the missing relationship and its consequences. |
| Does evidence match the candidate? | Source, build, inputs, procedures, runtime and configuration | Assess reuse or rerun. |
| Do results support every scoped claim? | Expected/actual observations, reviews and explicit missing work | Assign the uncovered claim. |
| Are anomalies and exclusions dispositioned? | Rationale, scope, owner and decision record | Keep the review open. |
| Is the next decision authorized? | Quality decision and separate release/validation references | Route to the identified authority. |
Useful evidence statuses are planned, executed, failed, partial, blocked, stale and supported for a stated claim. Record acceptance separately. In the fixture, the corrected conversion is supported by the recorded checks; no device acceptance status exists.
06 / Avoid misleading closure
Common IEC 62304 change-review mistakes
| Mistake | Consequence | Better record |
|---|---|---|
| Treat the newest publication as the adopted basis | Review against the wrong edition or recognition scope | Exact source identities and the project's adoption decision. |
| Translate FDA Basic/Enhanced into an IEC class | Apply an unsupported classification | Separate decisions, rationales and authorities. |
| Use old output as the only oracle | Preserve the old defect with precise tests | Expected behavior derived from the authorized requirement. |
| Show only formatted text | Miss state and unit errors such as unavailable becoming zero | Assert the whole observable output contract. |
| Add overlapping case counts | Create an inflated impression of breadth | Separate campaigns, domains and counting units. |
| Reuse a component pass after interface changes | Leave new assumptions unexamined | Impact rationale and renewed interface evidence. |
| Close the issue because a script exits zero | Confuse runner success with quality disposition | Failure, correction, rerun and authorized decision kept together. |
| Treat a trace link as a complete lifecycle record | Hide missing activity or an unexamined assertion | Inspect the linked work and its basis. |
Ask: what important claim would still be unsupported if every current row passed? The missing-as-zero mutant passes the integer-only campaign while exposing the missing-data failure.
07 / Keep the same evidence connected
Use Proof for a bounded change-evidence workflow
Proof's Software Intent Graph connects requirements with code, hazards, checks and decision records. Its change workflow retains the intended change, relevant evidence, issues and freshness information. For this reader task, those links can help assemble the same record shown above.
Start with one reviewed requirement and its actual authority. Connect affected code and tests, inspect candidate impact through recorded relationships, and attach results from the identified revision. Preserve unresolved behavior in issue records. Proof's public KnownIssue example illustrates expected/observed behavior and retained reproduction evidence; a KnownIssue is not automatically an accepted medical-device anomaly disposition.
Use the portal to inspect the project and MCP to retrieve context for an agent. Application checks run through the actual repository or CI and its permissions. Recorded links can miss an unrecorded dependency; trace percentages do not establish valid assertions or complete execution. Review impact and freshness against the real configuration.
Configure approval permissions to match the identified project authority. Proof's current initialization permits agent approvals at its internal assurance levels unless policy narrows them. Those product levels do not assign IEC software safety classes or medical-device decision authority. Current agent policy
This connected workflow does not supply a complete QMS, risk-management process, all lifecycle activities, tool qualification or regulatory submission package. No Proof audit or coverage measurement was performed in the display campaign.
For a smaller next step, use free Spec Lab to interpret one written accept/reject or missing-state rule and review proposed cases. It analyzes the interpreted rule; it does not execute application code, establish arbitrary numerical correctness, or provide medical/software conformity credit.
08 / Choose evidence for the claim
Which methods and workflow fit this change?
Choose the method by the question left open, then choose where to maintain its record. FDA's software guidance distinguishes reviews and analysis from testing and discusses unit, integration and system evidence. Definitions and testing guidance
| Question | Useful method | Decision boundary |
|---|---|---|
| Is the rule agreed and unambiguous? | Requirement review with the actual owner | A generated rule remains a proposal until handled by the project process. |
| Does a bounded calculation match that rule? | Literal oracle, exact arithmetic and domain enumeration | Accepted integer coverage does not cover missing or invalid states. |
| Can assertions detect a plausible mistake? | A selected, isolated fault or mutant | Detecting selected faults gives no universal adequacy score. |
| Did an interface or configuration change? | Integration checks on the identified target and configuration | Component evidence supports only its stated assumptions. |
| Does a dependency remain suitable? | Version/configuration review, known-problem assessment and focused regression | An inventory alone does not establish suitability. |
| Does the device serve its intended users and use? | Appropriate device-level validation under its plan | A correct conversion alone cannot settle intended-use suitability. |
For dependency review, FDA's OTS guidance offers prompts about version, configuration, limitations, problems, testing and maintenance. OTS terminology does not automatically determine an IEC SOUP classification. Record the project's actual assessment.
A controlled document works when the review is small and the owner keeps references current. Repository records with CI suit frequent changes when revisions and results can be retained together. A lifecycle platform or Proof can help people navigate larger sets of related records. Each approach still needs meaningful assertions, controlled authority and a clear decision history. This is a practical workflow comparison, not a vendor benchmark.
09 / Inspect the basis and reproduce the result
Primary sources and reproducible evidence
Sources by purpose
| Source | Use in this guide |
|---|---|
| IEC 62304 edition 1.1 | Published identity and lifecycle scope. |
| IEC stability FAQ and TC 62 strategic plan | Maintenance terminology and prospective standards work. |
| ISO 14971:2019 and IEC/TR 80002-1:2009 | Current risk-management publication and historical guidance with its original edition relationships. |
| FDA recognition 13-79 and standards-use guidance | Specific recognition and how recognition relates to revisions. |
| FDA device software guidance, June 2023 | Documentation levels, verification, changed versions and anomalies. |
| FDA OTS guidance, August 2023 and QMSR overview | Dependency documentation and current US quality-system context. |
| Python 3.14 Decimal documentation | Independent arithmetic cross-check semantics. |
| Proof graph, Changes, Agents and Spec Lab | First-party product workflow and its limits. |
The worksheet and procedure are original engineering advice. Determine applicable requirements from the actual adopted sources and responsible authority, rather than treating this source list as a conformity mapping.
Reproduce the captured contrast
Download the actual evidence kit, verify its ZIP checksum, and read README.md. From the extracted root, use Python 3.9+ and the standard library:
python3 -B reproduce.py --output rerun-results
Use a new output directory. The runner refuses to overwrite captured or nonempty output. It launches these four worker modes, each of which executes both campaigns:
python3 -B fixture/run_cases.py --implementation baseline
python3 -B fixture/run_cases.py --implementation corrected
python3 -B fixture/run_cases.py --implementation mutant_ties_even
python3 -B fixture/run_cases.py --implementation mutant_missing_as_zero
Expected child exits are 1, 0, 1, 1 in that order. Outer exit 0 reports successful reproduction of this contrast. No dependencies, network access, credentials or Proof CLI are needed. New run timestamps and runtime metadata can differ.
The captured worker commands use /opt/homebrew/opt/python@3.14/bin/python3.14 with the same arguments; PYTHONDONTWRITEBYTECODE=1 is recorded. The provenance preserves full command arrays and the final UTC timestamps. The capture remains CPython 3.14.7 even if the rolling documentation displays a later patch version.
Frozen identities and raw files
These SHA-256 values identify the captured source, inputs and runners:
1dbf9c38050fcae55ac1ba613fe76071aacd0fd8d72c4d95e6894d4182140a1e fixture/requirements.json
58ea3648eef23a31457d84c153a3c160f6b4c608135dfaa1126911ed01ca2c19 fixture/display.py
d0b3c9a892a4c5648834a26cc6943115c5a4dadb76872ed7afe9dfc1e6221899 fixture/cases.json
01d57d6816412b513f89dba1dc446d5d057f36ed4e429c94b852588e01e28f22 fixture/integer-domain-oracle.json
d8194d72e381cf79bdacd1129f0e7ba5acc6b419b399112b6ada0126e7113d4b fixture/run_cases.py
169c7162e98df3b10c74fc1a8a970f27de3b3038e58732909223878fa94f416d reproduce.py
| Inspectable artifact | Exact file |
|---|---|
| Worker and outer runner | fixture/run_cases.py, reproduce.py |
| Before and after | baseline-stdout.json, corrected-stdout.json |
| Isolated fault results | mutant_ties_even-stdout.json, mutant_missing_as_zero-stdout.json |
| Worker diagnostics | baseline-stderr.txt, corrected-stderr.txt, mutant_ties_even-stderr.txt, mutant_missing_as_zero-stderr.txt |
| Capture and kit checksums | captured/SHA256SUMS, SHA256SUMS |
The ZIP SHA-256 is ca2cd4b62b548bba32ec0046ab4ed22e5bf994d1940da75ac9794193942cea89. The kit also retains each worker's stderr file. Checksums identify bytes; they do not authenticate a reviewer, approve a requirement or establish what was loaded into a real product.
10 / Look beyond the local repair
Two additional change-review examples
These are proposed review situations, with no additional execution claimed.
1. The converter stays unchanged; the producer changes
A data-source update starts sending a last-known value after communication loss. The converter's integer tests can still pass because the value is well formed. Review the acquisition contract, age metadata, unavailable/stale distinction and every consumer of the displayed state. The requirement and risk owners must decide the intended behavior. Plan integration observations that distinguish fresh, missing and stale data; retain the local arithmetic result only for the unchanged arithmetic claim.
2. A display dependency changes version
A formatting-library update changes locale defaults or unit-label layout. Record the exact old/new dependency versions, configuration, documented changes and relevant known problems. Review which consumers use formatted text and which use the numeric value. Propose focused target checks for decimal formatting, labels and unavailable-state rendering. The dependency owner provides the assessment; quality review decides whether the evidence is sufficient. The earlier function output does not establish the new rendered display. FDA OTS documentation prompts
11 / Questions that affect the review
IEC 62304 change-evidence FAQs
Which IEC 62304 edition should my change record name?
Name the edition and amendments actually adopted by your project, with their approval reference. Keep publisher identity, national adoption and relevant regulator recognition separate. The published consolidated record discussed here is edition 1.1; a roadmap target does not change the project's basis. See the authority distinctions.
Does IEC 62304 cover medical-device validation and final release?
Its published scope excludes those device-level activities, including for software-only devices. This does not eliminate software release activities or the need for validation. Identify the plans and authorities that own them. IEC scope
Can I choose a software safety class from this example?
No. The fixture has no medical intended use or risk/class assessment. Record the actual project's classification and rationale through its responsible process. FDA documentation levels and Proof's internal assurance labels are different concepts.
Does 41/41 prove every input is handled correctly?
It covers the accepted integer domain in this captured contract. Missing input is outside it; selected invalid examples are not all possible invalid objects. The missing-as-zero mutant passes 41/41 while violating T07. Keep those claims separate.
Why not compare only with the previous version?
Agreement with a previous version can reveal regressions but can also preserve its mistakes. Here the previous rounding and missing-state outputs are wrong under the authored contract. An independently stated oracle makes intentional correction distinguishable from unintended change.
Can old verification evidence be reused?
Assess what changed in requirements, implementation, dependencies, target and configuration. Record the reuse rationale and its authority, or renew the affected evidence. Identical test names or a small diff do not establish that the assumptions still hold.
Can this worksheet or a Proof report establish conformity?
Neither can establish it alone. A report supports its identified claims and scope. The actual lifecycle work, risk decisions, applicable obligations, quality controls and release authority need their own evidence. Use the worksheet to make one change review inspectable.